If this is your first visit, be sure to check out the FAQ in HCMA Announcements. You may have to register before you can post: click the register link above to proceed. To start viewing messages, select the forum that you want to visit from the selection below. Your Participation in this message board is strictly voluntary. Information and comments on the message board do not necessarily reflect the feelings, opinions, or positions of the Hypertrophic Cardiomyopathy Association. At no time should participants to this board substitute information within for individual medical advice. The Hypertrophic Cardiomyopathy Association shall not be liable for any information provided herein. All participants in this board should conduct themselves in a professional and respectful manner. Failure to do so will result in suspension or termination. The moderators of the message board working with the HCMA will be responsible for notifying participants if they have violated the rules of conduct for the board. Moderators or HCMA staff may edit any post to ensure it conforms with the rules of the board or may delete it. This community is welcoming to all those with HCM we ask that you remember each user comes to the board with information and a point of view that may differ from that which you hold, respect is critical, please post respectfully. Thank you

Announcement

Collapse
No announcement yet.

More trouble for Guidant

Collapse

About the Author

Collapse

Lisa Salberg Find out more about Lisa Salberg
X
 
  • Filter
  • Time
  • Show
Clear All
new posts

  • More trouble for Guidant

    June 2, 2005
    Heart Device Sold Despite Flaw, Data Shows
    By BARRY MEIER
    When the Guidant Corporation told doctors last week that a popular implantable heart defibrillator had failed in a small number of cases because of an electrical flaw, it also said that it had fixed the flaw in devices produced after mid-2002.

    But now data provided by Guidant to a Minnesota hospital suggests that the company continued to sell the potentially flawed devices for months after it changed the way it made the device and had begun selling the new ones.

    The Guidant data, which the company provided to doctors at Abbott Northwestern Hospital in Minneapolis, indicates that from May to September 2002, nine patients there received implants of defibrillators that were made before April 2002.

    It was during that month that the company first changed its manufacturing process to eliminate the risk of short-circuiting. A senior company official has said that Guidant to date has not received reports of such failures in units made after April 2002.

    A defibrillator, which is implanted under the skin in the chest, acts to shock a chaotically beating heart back into normal rhythm. The device at issue is known as the Ventak Prizm 2 DR Model 1861, or alternatively as the Prizm 2 DR.

    In response to questions from The New York Times, Guidant said late yesterday that it had continued to sell units out of inventory - those made before the fixes - because it believed that the devices were reliable.

    In its statement, the company said: "After making these improvements, Guidant sold product manufactured before the improvements because the reliability data showed that the original PRIZM 2 DR, like the enhanced version, was a highly reliable life-saving product. Current data continues to support the reliability of this product."

    Last week, Guidant, which is based in Indianapolis, told doctors and patients for the first time that the defibrillator failed because of the short-circuiting problem seen in 26 known cases, including one recent death. About 37,000 such units were made before Guidant made the April 2002 fix. An estimated 24,000 people still use one of those units.

    The company said last week that the overall failure rate of the units made before and after the fix is similarly low. Last week, Guidant urged physicians not to replace the units because it said the risks of doing so outweighed any dangers the device may pose.

    However, the apparent fact that Guidant sold older devices with the potential flaw when improved ones were available could pose a new round of problems for the company. The Food and Drug Administration is reviewing how Guidant officials handled problems related to the device.

    In addition, two plaintiffs' law firms filed an action yesterday against Guidant in Federal District Court in Indianapolis seeking damages in connection with the company's handling of the device. Guidant has previously said that it did nothing wrong.

    The company's image is also still rebounding from an episode in which a former unit, Endovascular Technologies, pleaded guilty in 2003 to 10 felonies and agreed to pay $92.4 million to settle charges that it covered up deaths and other problems related to a device that treated weakness in the main abdominal artery.

    Officials of Guidant have insisted that they made all required reports concerning the Model 1861 to the F.D.A., including the information in individual reports from doctors and hospitals about any apparent problems with the device.

    According to the Guidant data, the nine patients at the Minnesota hospital represented 16 percent of the 57 patients there who received a Model 1861 made before Nov. 1, 2002. Those devices were implanted from December 2000 to January 2003. It was in November 2002 that Guidant made a second change to the unit.

    Some doctors said they would be dismayed if the company allowed them to implant a device with a known flaw that had been corrected in other units. Typically, a device like a defibrillator is not sold until a company sales representative delivers the unit to the surgical suite and it is implanted. Because their batteries drain, a defibrillator needs to be replaced every five or six years.

    Guidant is one of the nation's largest makers of medical devices, with $3.8 billion in sales last year, about half of them from implantable defibrillators. In December, Johnson & Johnson announced it planned to buy Guidant in a deal worth $25.4 billion. After the disclosure last week of the Model 1861's problems, Johnson & Johnson said it remained committed to the deal.

    "Guidant is an honorable company," said Dr. Douglas Zipes, a cardiology professor at Indiana University School of Medicine who is also a consultant to Medtronic, another defibrillator maker. "If they personally knew this device had the potential for failing, I can't imagine that they would go ahead and continue to sell it."

    The implant data at issue was provided by Guidant to doctors at Abbott Northwestern after the death in March of one of their patients, a 21-year-old college student with a genetic heart disease that put him at high risk of sudden death. In earlier interviews, those doctors, Dr. Robert G. Hauser and Dr. Barry J. Maron, expressed anger at Guidant for failing to alert them earlier to the unit's problems, saying they would have replaced it in their patient had they known.

    Dr. Zipes is the editor of a medical journal that recently published on its Web site a case study about the Guidant unit written by Dr. Hauser and Dr. Maron, though he was unaware of any issue related to inventory devices when interviewed last week.

    The student, Joshua Oukrop, died during a bicycling trip in Moab, Utah. After Guidant inspected his Model 1861 defibrillator, implanted in late 2001, it concluded that the device had malfunctioned because of the short-circuiting problem.

    While it is not certain when the device failed, the short circuit can occur when the device builds a charge to deliver the high-energy shock needed in emergencies.

    Dr. Hauser, who at one time headed a company, Cardiac Pacemakers, that has become part of Guidant, said he now faced telling some of those patients that they might have to undergo surgery to remove and replace a device they probably should not have received.

    "What am I supposed to tell these people?" asked Dr. Hauser.

    He agreed to provide a reporter with the implant data given to him by Guidant. Before sending it, personnel at Abbott Northwestern stripped out patient names and deleted portions of each device's serial number, which can also be traced to a patient.

    For each Model 1861 unit, the list gives the date of implant as well as a "use before" date, or the date beyond which the device can no longer be implanted because it has remained in inventory too long.

    An F.D.A. spokeswoman, Julie Zawisza, said that, according to the agency's information, the shelf life or "use before" date of the Guidant Model 1861 is 12 months after its battery is connected. Typically, that point marks a device's date of manufacture, and the unit goes into inventory a few weeks later, industry officials said.

    Based on that timeline, the Guidant implant data provided to Dr. Hauser suggests that nine devices with manufacturing dates before April 1, 2002, were implanted in patients at Abbott Northwestern after that time. For example, one device with a manufacturing date of August 8, 2001, was implanted in a patient on May 22, 2002, the data indicates.

    In addition, seven devices with manufacturing dates in February 2002, two months before Guidant started making fixes, were implanted in patients in July, August and September of that year, the data suggests.

    Guidant has said that in April 2002 it increased the spacing between a wire and a device component after determining that electricity could potentially arc between them and cause a short circuit.

    In November 2002, Guidant made another change by adding extra insulation around the component, which is called the backfill tube. The implant data from Abbott Northwestern shows that three patients received implants after November 2002 using devices with earlier apparent manufacturing dates.

    Both Dr. Hauser and Dr. Maron said that a senior Guidant executive, Dr Joseph M. Smith, had told them in recent weeks that the April 2002 fix had apparently solved the problem because the company had not received any reports of similar short circuits in units made since then. In a recent interview, Dr. Smith made the same statement to this reporter.

    In its statement yesterday, Guidant said: "The reliability data supports the conclusion that Prizm 2 DR was reliable when launched in 2000, when enhanced in 2002, and that it remains reliable today."

    Last week, in a notice sent to doctors about the potential electrical problem, Guidant referred only to the November 2002 change and stated that "devices manufactured after this date are not affected."

    Guidant has said it reported the November fix to the F.D.A. as part of an annual report to the agency, which it filed in August 2003. However, the company also said in response to an earlier inquiry from The Times that it did not report the April 2002 changes because it decided that F.D.A. regulations did not require the reporting of those "manufacturing enhancements."

    The company said that the changes did not have to be reported because they only "added clarifications to existing manufacturing instructions and improved process consistency," rather than reflecting changes to approved F.D.A. manufacturing techniques.
    Knowledge is power ... Stay informed!
    YOU can make a difference - all you have to do is try!

    Dx age 12 current age 46 and counting!
    lost: 5 family members to HCM (SCD, Stroke, CHF)
    Others diagnosed living with HCM (or gene +) include - daughter, niece, nephew, cousin, sister and many many friends!
    Therapy - ICD (implanted 97, 01, 04 and 11, medication
    Currently not obstructed
    Complications - unnecessary pacemaker and stroke (unrelated to each other)

  • #2
    The Flaws in a Heart Device (2 Letters)
    E-Mail This
    Printer-Friendly


    Published: May 31, 2005
    To the Editor:

    Re "Maker of Heart Device Kept Flaw From Doctors" (front page, May 24):

    I was appalled to learn that manufacturers of medical devices do not always disclose information about flaws to doctors and patients. It is unconscionable that companies like Guidant would withhold information that could prove fatal to high-risk patients because of fears about their reputation.

    Manufacturers of such devices must be held to higher ethical standards than manufacturers of other products.

    Hannah Hahn
    New York, May 25, 2005


    To the Editor:

    According to your article on Guidant (front page, May 24), "a top Guidant executive, Dr. Joseph M. Smith, said that the company had not seen a compelling reason to issue an alert to physicians about the defibrillators because the failure rate was very low and replacing the devices might pose greater patient risks."

    This is arrogant and demonstrates an ignorance of contemporary bioethics. Corporate paternalism is inappropriate. A decision about whether to replace a possibly defective defibrillator should be made jointly by the doctor and the patient after consideration of all available data.

    Herbert Rakatansky, M.D.
    Providence, R.I., May 25, 2005
    The writer is a clinical professor of medicine at Brown University.
    Knowledge is power ... Stay informed!
    YOU can make a difference - all you have to do is try!

    Dx age 12 current age 46 and counting!
    lost: 5 family members to HCM (SCD, Stroke, CHF)
    Others diagnosed living with HCM (or gene +) include - daughter, niece, nephew, cousin, sister and many many friends!
    Therapy - ICD (implanted 97, 01, 04 and 11, medication
    Currently not obstructed
    Complications - unnecessary pacemaker and stroke (unrelated to each other)

    Comment


    • #3
      Thanks for posting this info Lisa. My EP is based at RI Hospital and I believe all the docs in and around the Hospital and Brown University are more then a little upset and I echo their sentiment.

      I did have a conversation with a local reporter whose paper is going to do a story. The paper contacted me and would like to incorporate the story of my HCM along with this . The reporter is away presently but should be contacting me soon. After the conference you can give me some pointers as to important facts to give to them. These articles would be usefull in tying HCM in with the importance of trust issues for safe working life saving devises.

      I think the unfortunate circumstances that led to our learning of this devise failure gives all of us an opportunity to do some education perhaps in our local areas about HCM ( because many of us with HCM rely on the proper function of these devises) and more importantly we rely on and trust that our best interests and wellfare will be first and foremost when it comes to disclosure of information.

      I will keep you informed.

      Pam
      Dx @ 47 with HOCM & HF:11/00
      Guidant ICD:Mar.01, Recalled/replaced:6/05 w/ Medtronic device
      Lead failure,replaced 12/06.
      SF lead recall:07,extracted leads and new device 2012
      [email protected] Tufts, Boston:10/5/03; age 50. ( [email protected] 240 mmHg ++)
      Paroxysmal A-Fib: 06-07,2010 controlled w/sotalol dosing
      Genetic mutation 4/09, mother(d), brother, son, gene+
      Mother of 3, grandma of 3:Tim,27,Sarah,33w/6 y/o old Sophia, 5 y/o Jack, Laura 34, w/ 5 y/o old Benjamin

      Comment


      • #4
        I was implanted on march 26th of 2003 so that explains why she said (at the company) that i was included and that i had the device in question and that i needed to go to my Dr. to have it checked, so i guess i will get zapped then being that seems to be the flaw, that is a scary thought, well it has to be done so i have an appt on the 6th to get it checked

        Shirley
        Diagnosed 2003
        Myectomy 2-23-2004
        Husband: Ken
        Son: John diagnosed 2004
        Daughter: Janet (free of HCM)

        Grandchildren: Drew 15,Aaron 13,Karen 9,Connor 9

        Comment

        Today's Birthdays

        Collapse

        Working...
        X